Another thing to bear in mind with your apparent low readings on Libre
@Garlo is the algorithm that the device is using to try to reduce the lag between interstitial fluid and capillary glucose from 10-15 minutes to approximately 5 minutes.
I am sure this is covered in the ‘CGM precautions’ post that
@helli linked to, but effectively the algorithm tries to look ahead based on current value and trajectory and estimate where glucose values are likely to be in the next 5-10 minutes. This is generally very accurate and helpful, but there
can be occasions where glucose levels are changing rapidly (and particularly where they are rising/falling and then abruptly change direction) where Libre scans ‘overshoot’, and show a reading that you never actually reach, and then later iron that out with their 5 minute averages plotted as the sensor trace.
You might also find this study reassuring, which shows glucose variation recorded in healthy individuals (without diabetes) via CGM.