Docb
Moderator
- Relationship to Diabetes
- Type 2
In my quest to help people get a perspective on test results and what you can get from them, I have done some more systematic tests.
They follow on from the tests I did some time ago when I jabbed all eight fingers and two thumbs in quick succession to see by how much they varied and reported on here. There I came to the conclusion that the number after the decimal point was a bit meaningless and that you could easily get a variation of +/- 1 around the average reading. I suggested that was as much to do with the variability in blood glucose around the body as "inaccuracy" in the meter.
This time I decided to see how many tests I could get from a single drop of blood. Might go some way to differentiating between sample variability and reader variability. So, bodged the finger to get a decent drop of blood, got a reading, replaced the test strip and tried again. On two occasions I got two readings and on the third occasion I got three. The three attempts were at different times - bedtime last night, one hour and a half after breakfast this morning and then again at 10:00 AM. I was aiming to get my blood glucose at different levels to see if that effected things.
The results were... Test 1: 6.2, 6.8
Test 2: 9.7, 10.1
Test 3: 7.6, 7.3, 7.2.
A quick eyeballing of the numbers suggests you cannot expect a reproducibility much better than half a unit and it is more sensible to assume a reproducibility of 1 unit.
Thought some of the number nurds might be interested. For everybody else, be thankful for the fact that there is kit around which can get some sort of instantaneous reading of blood glucose but don't get too wound up by changes of a unit or two. You can get that by repeating the test on the same drop of blood. The kit is very good, but not that good.
They follow on from the tests I did some time ago when I jabbed all eight fingers and two thumbs in quick succession to see by how much they varied and reported on here. There I came to the conclusion that the number after the decimal point was a bit meaningless and that you could easily get a variation of +/- 1 around the average reading. I suggested that was as much to do with the variability in blood glucose around the body as "inaccuracy" in the meter.
This time I decided to see how many tests I could get from a single drop of blood. Might go some way to differentiating between sample variability and reader variability. So, bodged the finger to get a decent drop of blood, got a reading, replaced the test strip and tried again. On two occasions I got two readings and on the third occasion I got three. The three attempts were at different times - bedtime last night, one hour and a half after breakfast this morning and then again at 10:00 AM. I was aiming to get my blood glucose at different levels to see if that effected things.
The results were... Test 1: 6.2, 6.8
Test 2: 9.7, 10.1
Test 3: 7.6, 7.3, 7.2.
A quick eyeballing of the numbers suggests you cannot expect a reproducibility much better than half a unit and it is more sensible to assume a reproducibility of 1 unit.
Thought some of the number nurds might be interested. For everybody else, be thankful for the fact that there is kit around which can get some sort of instantaneous reading of blood glucose but don't get too wound up by changes of a unit or two. You can get that by repeating the test on the same drop of blood. The kit is very good, but not that good.