Abbott (ABT) today announced it is initiating a voluntary recall of FreeStyle InsuLinx? Blood Glucose Meters in the United States.
The company has determined that at extremely high blood glucose levels of 1024 mg/dL and above, the FreeStyle InsuLinx Meter will display and store in memory an incorrect test result that is 1024 mg/dL below the measured result. For example, at a blood glucose value of 1066 mg/dL, the meter will display and store a value of 42 mg/dL (1066 mg/dL ? 1024 mg/dL = 42 mg/dL). No other Abbott blood glucose meters are impacted by this issue.
http://finance.yahoo.com/news/abbott-initiates-voluntary-recall-freestyle-110000168.html
(1024 mg/dl = 56.9 mmol/l 😱)
The company has determined that at extremely high blood glucose levels of 1024 mg/dL and above, the FreeStyle InsuLinx Meter will display and store in memory an incorrect test result that is 1024 mg/dL below the measured result. For example, at a blood glucose value of 1066 mg/dL, the meter will display and store a value of 42 mg/dL (1066 mg/dL ? 1024 mg/dL = 42 mg/dL). No other Abbott blood glucose meters are impacted by this issue.